ompliance voltage is the voltage available at the counter electrode that can be used to force current to flow and still maintain control of the working electrode voltage.

How to Speak Potentiostat Compliance

Dr. Bobs Speaking Potentiosta ComplianceWhat’s important in the selection of a potentiostat? That depends greatly on its intended use. This page addresses an often misunderstood term.

Compliance voltagePotentiostat: Compliance Voltage

Compliance voltage is the voltage available at the counter electrode that can be used to force current to flow and still maintain control of the working electrode voltage. The compliance should be specified in conjunction with a current value: “20V @ 250mA” or “20V @ full rated current”. An amplifier’s output voltage decreases at higher currents due to the output impedance of the amplifier.

[…]

Gamry Interface - Reference Potentiostats

Potentiostat Fundamentals

Introduction

A potentiostat is an electronic instrument that controls the voltage difference between a Working Electrode and a Reference Electrode. Both electrodes are contained in an electrochemical cell. The potentiostat implements this control by injecting current into the cell through an Auxiliary or Counter electrode.

In almost all applications, the potentiostat measures the current flow between the Working and Counter electrodes. The controlled variable in a potentiostat is the cell potential and the measured variable is the cell current.

Gamry Potentiostat

Prerequisites

This Application Note may be difficult to follow unless you have some familiarity with electrical terms such as voltage, current, resistance, frequency, and capacitance. If you feel your knowledge in this area is lacking, we suggest review of a very basic electronics or physics book.

Electrodes

A potentiostat requires an electrochemical cell with three electrodes as shown below. W/WS denote the working and working sense. R denotes the reference electrode and C denotes the counter electrode.

Potentiosta: Electrochemical cell-3 electrodes […]

There is a description of potentiostat stability (written by DK Roe) in the Kissinger & Heineman book (

Converting Potentials to Another Reference Electrode

Often we find a potential listed in the literature quoted against a different reference electrode than the one we favor, or we would like to convert the potential to a more commonly used reference electrode for publication. A student emailed me: “My experiments involve measuring the redox potential relative to a saturated Ag/AgCl reference electrode. Read more about Converting Potentials to Another Reference Electrode[…]

Gamry Reference Electrodes

Potentials of Common Reference Electrodes

The following tables give the potentials of several commonly used reference electrodes. Various filling solutions are listed where data was available. Note that the nature and concentration of the filling solution can dramatically change the potential! When you look through the scientific literature, be sure the author has specified the filling solution! If you are Read more about Potentials of Common Reference Electrodes[…]

There is a description of potentiostat stability (written by DK Roe) in the Kissinger & Heineman book (

Potentiostat Stability

Contact Dr Bobs Buzz - Electrochemical Methods

A while ago I received an email from an electrochemist who lamented:

“We have some problems with the 173, which we still prefer to use occasionally  because of its analog nature. … (The) potentiostat goes into oscillations.”

Although the M173 has a reputation for ‘stability’ it has always had problems with oscillation! These problems tend to be most troublesome when the more sensitive current ranges are used and when the cell capacitance is large. PAR had a ‘noise filter’ in their catalog for a long time that was really a ‘stability aid’ more than a ‘noise filter.’ It consisted of a capacitor that was placed between the counter electrode lead and the input jack of the M178 electrometer. This acts as a shunt for the higher frequency, oscillation-producing signals. A capacitor value of 0.01 µF is a good place to start. I think this stands the best chance of stabilizing your system.

[…]

This website uses cookies to ensure you get the best experience on our website. More information

The cookie settings on this website are set to "allow cookies" to give you the best browsing experience possible. If you continue to use this website without changing your cookie settings or you click "Accept" below then you are consenting to this. Please review our Privacy Policy

Close